请教一下, 在1500V,或3000V耐压测试中,为什么漏电流在耐压仪里设定是10mA? 在IEC/UL60950里有规定吗?
是否是这一节: 6.1
For equipment powered from an AC MAINS SUPPLY, the test is performed with a voltage equal to
the RATED VOLTAGE of the equipment or to the upper voltage of the RATED VOLTAGE RANGE. For
equipment powered from a DC MAINS SUPPLY, the test is performed with a voltage equal to the
highest nominal voltage of the AC MAINS SUPPLY in the region where the equipment is to be
used, for example, 230 V for Europe or 120 V for North America.
The current flowing in the test circuit of Figure 6A shall not exceed 10 mA.
但我认为这一章是对通迅网络而言的. 不是对整机和变压器而言的. |
|